This forum exists for the purposes for discussing service monitors (This includes but is not limited to Motorola, HP, Aeroflex, GD, etc). Additional topics allowed include test procedures, interpretation of test results, where to find information about specific tests, antenna VSWR, return loss testing, duplexer and filter alignment, etc.

Moderator: Queue Moderator

Post Reply
User avatar
bradlington
1 Warning for RSS/CPS Wanted/For Sale/Links
Posts: 155
Joined: Sun Jan 05, 2003 10:34 am

HP8920A problem locking up

Post by bradlington »

HP8920A

I have the following problem with the HP8920A.

Over a short period the machine did the following:

Lost seconf audio generator(normally used to generate CTCSS etc)
Lost the audio on speaker of unit.
Unit would work for short while and lock up ---no chaning on menu .
Continuos ERROR 0040 after boot-up.

I did the following :

Removed each all cards in unit and cleaned with brush and compressed air.
Removed and reseated ALL Eproms.

Now the unit powers up and displays"ALL TESTS PASSED"--but still have the following issue:
All works but ;
second generator still faulty
no audio
deviation seems to be limited to max reading of 1,8 Khz


This is the strange thing : When I adjust the squelch from fully counter clockwise to clockwise-as soon as the squelch is triggered- the screen locks up and only way to resore is to power rest unit .

Any similar issues.

Thanks

Bradley