Page 1 of 1
IFR 1500S
Posted: Tue Jan 31, 2012 5:15 pm
by Highvoltage2010
My IFR 1500S has develpoed a problem with the spectrum analyzer. The sensitivity and trace are not right. Other functions work good including scope etc.. Self check works OK, no errors.
Has anyone run across this before. recommendations on reasonable repair?
thanks
Re: IFR 1500S
Posted: Wed Feb 01, 2012 4:32 am
by Wowbagger
What do you mean the trace "[isn't] right"? Too low, doesn't show the carrier, looks like Hamilton J. Frog, what?
Have you checked your sensitivity - if you generated a -120dBm FM signal out the duplex port and loop that to the receiver, what happens?
Re: IFR 1500S
Posted: Mon Feb 06, 2012 2:49 pm
by jry
There are a number of modules dedicated to the spectrum analyzer. You may want to check out the maintenance manual on the repeater-builders site.
There are full schematics , troubleshooting and alignment information in the 7MB version.
You may need some other supporting discrete test equipment to take this on ( sig gen, scope and spectrum analyzer <1Ghz )
The 1500 can look daunting but if you take it slow and read through the above information you will find it very servicable with the above tools.