Calibrate COM120B

Posted: Thu Jul 15, 2010 6:08 pm
by midstate33
OK, I'm satisfied I have my COM120B calibrated, how do I get rid of the "Out of Calibration" message displayed each time
its turned on, if the procedure is in the manual I'm over looking it, any help is appreciated. Tnx Fred/W8WAR

Re: Calibrate COM120B

Posted: Fri Jul 16, 2010 3:51 am
by Wowbagger
If you modify any calibration parameter, the date of cal will be reset. So if you have actually changed any of the cals from the software, that message should have disappeared.

If all you've done is verify the unit is still in cal, you just need to go into the cal screens, open an edit on any one of the calibration values, and close it - that will reset the date and clear the message for a year.