R2670 FDMA BER 1011 test patterns
Posted: Thu Feb 17, 2011 1:38 pm
Ran into a reproducible symptom with several of these monitors we have.
Wondering if others have experienced the same issue.
When in GENERATE mode, and running the 1011 test pattern for say a
GTR8000 receiver BER test. The level will be adjusted for an approx 5%
BER figure.
Letting it just sit there idle, the value will suddenly begin to creep up.
To say 7%, 12%, 22% - then back down to 3% and resettle at the original
approx 5% value.
After about a minute or two, it'll repeat the same cycle.
It's definitely caused by the 2670. When we used our other Aeroflex 3920,
the BER figure was rock solid.
Seems like the 2670's have some kind of design flaw if I didn't know
better.
If anyone has time, try the 1011 pattern and measure using the BER
screen in CPS. Curious if other people observe the same symptoms.
Wondering if others have experienced the same issue.
When in GENERATE mode, and running the 1011 test pattern for say a
GTR8000 receiver BER test. The level will be adjusted for an approx 5%
BER figure.
Letting it just sit there idle, the value will suddenly begin to creep up.
To say 7%, 12%, 22% - then back down to 3% and resettle at the original
approx 5% value.
After about a minute or two, it'll repeat the same cycle.
It's definitely caused by the 2670. When we used our other Aeroflex 3920,
the BER figure was rock solid.
Seems like the 2670's have some kind of design flaw if I didn't know
better.
If anyone has time, try the 1011 pattern and measure using the BER
screen in CPS. Curious if other people observe the same symptoms.